The fact that both profiles produced good looking prints is obviously a good sign. If you are using Canon own paper and ink how do the profiles compare to Canons own?Hi, first time posting here.
I have 3D-printed the two-part scanning jig and I have used it with the 480 patch argyll target provided by pharmacist. I scanned the target with my ColorMunki Design device and obtained very low dE values (max of 1.3 or something like that). I followed the procedure describe here. However, the white point of the profile (as shown on iccview.de for instance) is signficantly lower than for a profile obtained with ccStudio on the same paper (Canon SG-201 paper, see below).
Both profiles provide good-looking prints when tested on popular printer test images (for instance the "Outback print image") but I still worry about the white point.
View attachment 16828
Using the spotread command I noticed that argyll measures a LAB value of about -93 / 0 / -3 when the instrument is pressed on the paper (much closer to the value found by the xrite Software), but that value drops to about -88 / 0 / -3 when it is put on the scanning jig - consistent with my profile.
I believe the ColorMunki must be pressed against the paper but when it sits on the scanning jig there's an almost 2 mm gap between the hole of the instrument and the paper. That likely causes some light diffusion, thus the lower L value? Actually when I press the instrument on the paper and tilt it only very slightly (like 1mm) this is enough to have a lower L reading...
Can you confirm this on your side?
I should mention that my jig is made of white PETG but painting the inside of the slit in black did not change anything.
I can neither confirm or deny the differences as I have not tried this method properly or made the comparisons you have, however your findings seem to be in agreement with Argylls author.
Quote from Graeme Gill (Developer ArgylCMS)
The Munki/i1Studio isn't designed to be spaced away from the paper - the geometry of the illuminant doesn't allow for that.
So it really should be sliding on the paper itself. A "gotcha" is that the instrument has two small rubber pads that
are designed to stop it slipping when taking individual measurements, and these impede smooth sliding. If
you are mainly doing strip reading you could remove these, or if you leave them they will wear down in time anyway
At the end of the day the differences may be there but insignificant for the type of paper being used and perfectly acceptable without being technically correct accepting that YMMV
FWIW dE difference is just over 3